Beep
Beep
P0623-A
Figure 19: Ground Balance Procedure
NOTE
Movement of the coil during the entire Ground Balance procedure
should be slow, continuous and smooth and each down and up
movement should take 3 to 4 seconds.
If the Ground Balance OK tone is not heard within
30 seconds of starting the procedure, release the Ground Balance
button and repeat this procedure.
If there is metal in the ground under the coil whilst Ground
Balancing, the detector will not Ground Balance correctly. Move the
detector and repeat the Ground Balance over ground that is free of
any metallic objects.
After the Ground Balance procedure is completed the detector will
automatically cancel interference from the ground under the coil. If
ground conditions change (changing mineralisation in the ground)
this procedure may need to be repeated.